PH ? - Liquid test kit vs ph meter

Here's the stats:

RO/DI water, KH by potassium carbonate/snail shells, GH by Seachem Replenish. (switching to GLA GH booster in the near future.)

TDS out of the RO/DI is 0ppm, checked with two TDS meters.

KH is 3
GH is 4-5
No CO2 injection, yet.
Two india almond leaves in tank.

Now onto my question.

When I check the pH with my calibrated meter I get a pH of ~6-6.2, doesn't mater what time of day.

When I check the pH with the API kit I get ~7.0-7.2.

I check both samples together.

What I do:
I keep the meter in the standard solution which is pH 7.0, I kick on the meter before pulling it out of solution and it always reads 7.0.
I then rinse the meter and API glass tube in RO/DI water, then rinse in aquarium water. Dump all that out.

I then extract water from the tank, put it in the glass container rinsed by RO/DI-aquarium water, stick the pH meter in, swirl water with it, and then take the 5ml of water out for the API test.

So I'm watching my meter as I put drops into the tube and shake.

The test drops are of recent manufacture and don't expire for a long long time.

Why are they different and which one is it?

Thanks in advance!


To view links or images in signatures your post count must be 10 or greater. You currently have 0 posts.
Calestus is offline  
Sponsored Links
Advertisement
 
post #2 of 6 (permalink) Old 10-16-2015, 05:12 AM
Planted Tank Guru
 
PTrader: (84/100%)
Join Date: Dec 2005
Location: Sacramento, CA
Posts: 21,015
Measuring pH with accuracy looks easy, but it isn't. If you use the test kit, where you use the color of the water plus reagent to tell you what the pH is, it is very hard to judge the color accurately. You will be looking at a printed chart, with unknown color accuracy, as the standard, and comparing colored water in a round cross section tube to that. But, the color of the water will look different depending on how you view it. Getting a +/-0.2 pH accuracy almost takes a miracle with that test kit.

If you use a pH probe you have to calibrate the probe with standard pH solutions often, and do it correctly. The probe calibration can drift with time and usage, so you need to recalibrate it often. If you do that well you may be able to get a +/-0.1 pH accuracy.

I would trust the probe reading over the liquid test kit reading almost every time. But, I'm surprised that you got that big a difference. Something isn't right.

Hoppy
Hoppy is offline  
post #3 of 6 (permalink) Old 10-16-2015, 06:40 AM
Algae Grower
 
PTrader: (1/100%)
Join Date: May 2015
Location: Arkansas
Posts: 114
Can you run the kit on the 7.0 standard and see what it says? Like Hoppy I'd be inclined to trust the probe, especially since you say it reads the standard as 7.
Vidikron is offline  
 
post #4 of 6 (permalink) Old 10-16-2015, 10:41 AM
Wannabe Guru
 
jrill's Avatar
 
PTrader: (4/100%)
Join Date: Nov 2013
Location: Near Ashville, North Carolina
Posts: 1,866
You need to calibrate your probe separate from your storage solution.
jrill is offline  
post #5 of 6 (permalink) Old 10-16-2015, 05:41 PM
Algae Grower
 
PTrader: (0/0%)
Join Date: Sep 2015
Location: Pennsylvania
Posts: 39
I had never thought to check the pH of the storage solution before reading this thread. I just checked, the storage solution I use is ~7.3.

These are what I'm using:
http://smile.amazon.com/gp/product/B002NX0W2I
http://smile.amazon.com/gp/product/B001D0CKYK
Not saying it's the best stuff to use, just what I happen to be using. And, I really don't know why you can't use calibration solution for storage, I just know they say you shouldn't long term.

My pH probe came with poorly written basic use instructions. Here is a more in depth write-up I found, Care and Storage of a pH Electrode.

FWIW - When I've checked the probe versus my API kit, they were both relatively close. I do believe the pH probe more so than telling the difference between shades of blue/green.
SL Dan is offline  
post #6 of 6 (permalink) Old 10-17-2015, 04:45 AM Thread Starter
Algae Grower
 
Calestus's Avatar
 
PTrader: (2/100%)
Join Date: Oct 2015
Location: Utah
Posts: 135
So I recalibrated my meter with fresh standard and it was ~.2 off to my old standard.

I then took three samples of tank water and one sample of my RO/DI water and did the API test on it.

They all read pretty much the same...now. I'm not sure what happened or why there was such a discrepancy before. I tested a few times before I posted and got the same ~1PH difference over several days.

I haven't done a water change or added anything since.

*shrugs*

Thanks for the replies you all!
Calestus is offline  
Reply

Tags
None

Quick Reply
Message:
Options

Register Now



In order to be able to post messages on the The Planted Tank Forum forums, you must first register.
Please enter your desired user name, your email address and other required details in the form below.

User Name:
Password
Please enter a password for your user account. Note that passwords are case-sensitive.

Password:


Confirm Password:
Email Address
Please enter a valid email address for yourself.

Email Address:
OR

Log-in










Thread Tools
Show Printable Version Show Printable Version
Email this Page Email this Page
Display Modes
Linear Mode Linear Mode



Posting Rules  
You may post new threads
You may post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On

 
For the best viewing experience please update your browser to Google Chrome